Snake venoms typically are proteins in the size range 15-20 KDa. That means they don't differ much by size, so SEC won't give you much resolution. Also, people usually want to retain the enzymatic activities of the proteins, which rules out denaturing modes like RPC. That leaves HIC (hydrophobic interaction chromatography, NOT HILIC [hydrophilic interaction chromatography]) and ion-exchange. Francis Markland's group at USC has used both modes to purify various snake venom proteins to homogeneity. Here are some references:
1) A.D. Retzios and F.S. Markland, Prot. Exp. Purif. 1 (1990) 33.
2) A.D. Retzios and F.S. Markland, Jr., Biochemistry 31 (1992) 4547.
3) M. Trikha, W.E. Rote, P.J. Manley, B.R. Lucchesi, and F.S. Markland, Thrombosis Res. 73 (1994) 39.
4) S.L. Loayza, M. Trikha, F.S. Markland, P. Riquelme, and J. Kuo, J. Chromatogr. B, 662 (1994) 227.
They used our company's HIC columns and ion-exchange columns from SynChrom. The latter are pretty much gone from the market, but we can supply you with good alternatives for those as well. Contact us offlist if you want to follow up.
Andy Alpert
PolyLC Inc.
(410) 992-5400
info@polylc.com