Method sensitivity issue-urgent

Posted: Sun Jan 23, 2011 8:26 pm
by darasc0211
Hi,

My sample is a super low dose vitamin D5 and my method has a sensitivity issue with UV detector. I've tried using 100ul injection volume, increased smp concentraion, and used the wavelength that has the maximum absorption, but my peak area is only 5. Does anyone know if Vitamin D5 have fluorephore? or can be derived so that fluorosence detector can be used? Or any other idea at all?

Thanks.

Re: Method sensitivity issue-urgent

Posted: Sun Jan 23, 2011 11:31 pm
by tom jupille
Sensitivity boils down to signal/noise. So:
How tall is your peak (in absorbance units)?
How much noise are you seeing (also in absorbance uits)?

Re: Method sensitivity issue-urgent

Posted: Mon Jan 24, 2011 2:42 am
by carls
Is a longer pathlength flow cell available for your UV detector?

Re: Method sensitivity issue-urgent

Posted: Mon Jan 24, 2011 12:11 pm
by Mattias
One also very important question is how much sample you have. If you have "unlimited" amount you can try for an increased injection volume (1 ml?) in combination with precolumn trappning of your analyte. You will need some new hardware though (new syringe, loop, column switch).

Re: Method sensitivity issue-urgent

Posted: Tue Feb 01, 2011 2:56 am
by darasc0211
Due to the budget, we don't have the choice of changing a bigger UV cell pathway like 2cm, or larger loop, I only have 25mg of drug for the whole development. And that's very stressful! I am pulling my hair.

Re: Method sensitivity issue-urgent

Posted: Tue Feb 01, 2011 3:42 am
by tom jupille
So approach the problem by decreasing your baseline noise. That doesn't cost you any sample.

Re: Method sensitivity issue-urgent

Posted: Tue Feb 01, 2011 4:46 am
by darasc0211
So approach the problem by decreasing your baseline noise. That doesn't cost you any sample.
Hi,

How shall I decrease baseline noise? Could you give me a list that I could try? I am already using a 3um column. Thanks.

Re: Method sensitivity issue-urgent

Posted: Tue Feb 01, 2011 5:38 am
by bisnettrj2
See these two articles from John Dolan on decreasing baseline noise and increasing the signal-to-noise ratio.

http://chromatographyonline.findanalyti ... rticle.pdf

http://chromatographyonline.findanalyti ... ?id=660930