Long time reader, first time poster. I hope that you can help me with a problem that I have been having for several days now. We do a lot of peptide mapping via LC-UV (Agilent 1100 and 1200 systems). We use water/acetonitrile gradients with 0.1% TFA in water and 0.085% TFA in acetonitrile. The problem we are having with several of our system is that we get baseline noise during the early portion of the gradient, and then this noise goes away as the gradient progresses. I do not have a chromatogram to show you - I will try to attach one next week when I am at work.
For example, a typical gradient may go from 3% acetontrile to 60% over 45min. We will see a baseline oscillation, which may range up to 10mAU in amplitude, during the beginning of the run. As the gradient progresses, this noise will get smaller and smaller. Eventually, at ~20-30% acetonitrile, the noise will be negligible. However, the size the oscillations in baseline during the early portion of the run can obscure our ability to analyze the smaller, early-eluting peptides. By the way, we see this oscilaltion at 210 and 214nm, but it goes down as we increase the wavelength. Using refernce wavelengths does not seem to help.
We see this on different HPLC systems, and if we switch the water an acetontrile positions, we still see the same effect. The pump pressure seems fine - there is no significant ripple in pressure, and the system spass leak and detector tests.
I have a feeling that this has to do with build-up of TFA int eh systems - has anyone any similar experiences? Is there a method for cleaning residual TFA out of a system? This is my best hypothesis becuase we went to a system which had never been used for TFA, and within 24hrs we saw the same absorbance oscillations occurring.
Please help. I will try to post a chromatogram on Monday if I do not have an answer by then.