Having worked for a company that did not encourage publishing by us hoi polloi, I never published the following tidbit, or even checked whether it was already published. For those working in method development, it could speed the optimization of mobile phase pH considerably, as it did for me.
Anyone who has done buffer calculations is probably aware of the "sigmoid" curve titration plot (volume reagent added versus pH). I don't believe it is commonly appreciated that this plot need not be sigmoid, but can be linear. That is to say, within a specified region of pH, it is possible to have a linear relationship between titrant and pH.
More to the point with respect to HPLC mobile phases, it is possible to have two stocks of buffer reagents, and linear change from one to the other will, within a wide range, give a linear and predictable buffer pH. (Note always that when buffer is combined with organic solvent, the apparent pH as measured by the glass electrode will shift. I am speaking of the pH before addition of any organic solvent.)
The "magic" involves in choosing a series of pKa's that are approximately 1.5 pH units apart. (I have no intention of proving this here. I have and can prove it mathematically, but more importantly, I have demonstrated it experimentally. What is important is that this is a calculated result that has been confirmed experimentally, and those calculations can be confirmed by anyone with a slide rule or computer, and those experimental confirmations can be reproduced by anyone who has an analytical balance and a pH meter.)
Curiously, citric acid is a perfect polyacid for this use, as its three pKa's differ by almost exactly 1.5 pH units. (Note that some published values for the pKa's of citric acid include one wrong value -- I forget which. I discovered this when I couldn't confirm the published pKa's, and later found other published values that agreed with my calculations.)
Hence, if you make, for example, 100 mM trisodium citrate solution, and 100 mM citric acid solution, you will find that across the midrange of possible mixtures, the pH will vary linearly with the volume fraction of the former in the mixture (or inverse linearly with the volume of the latter). At the extremes, the linearity is lost, but this still gives a wide range of pH control BY VOLUMETRIC MEANS.
(Note that citric acid can be corrosive to stainless steel and other metals. If using citric acid in HPLC systems, purge after use to prevent the mobile phase from attacking the metal. This is also true of other chelating agents and of chloride ion.)
There is no reason that a single polyacid is needed. Any series of acids or bases (so long as they don't react, precipitate, etc.) can be employed for this purpose. So, for example, one could choose three bases with pKa's of 7.5, 9, & 10.5 and create a two-component buffer system that would range linearly from about (I'm guessing here) pH 6.5 to pH 11.5. You can calibrate this if you wish, so you know the relationship between volume and pH, but it isn't necessary to do so, as it's simpler just to measure the pH of the buffer system that works best for you. This is to say that the importance of linear variation in pH is not that it's predictable, but that it is gradual and reproducible, so you can come back later and determine the exact pH you found to work best on the basis of the volume mixture you used during the chromatography.
So, how is this useful for method development? Well, ternary mixers are common. Consider reversed-phase HPLC (e.g., C18 column). Devote A to methanol, B to acetonitrile, C to 100 mM citric acid and D to 100 mM trisodium citrate. Start with some level of A, with the balance 50:50 C & D, and vary the level of A across a few chromatograms. Next use something less of B (as acetontrile is a "stronger" solvent than methanol), with the balance 50:50 C & D, and vary the level of B across a few chromatograms. A few such tests will tell what the organic can do for you, and might suggest whether a mixture of methanol and acetonitrile will be beneficial (possible, but rare in my experience). Now start with some results that look promising and vary the C-to-D ratio, while leaving the organic alone. This alters pH only, and in a linear way. If pH matters in your separation (which it may if you're dealing with acids and bases, but probably won't if you're not), then you'll see changes of interest.
Find some "optimal" results, but don't spend a lot of time on them because the next thing you'll do is to change the buffer system entirely. (As mentioned, citric acid is a poor choice as buffer due to corrosion effects. Or, if you're using a mixture of other acids or bases, you might not need all these in the final mobile phase.) Determine the ratio of C to D you prefer, pump this w/o organic until you've collected maybe 50 mL, then take it to a pH meter and read the actual pH.
Once you know the pH you need, choose a buffer that will provide this pH -- namely, one that has a pKa within one pH unit of the desired pH. (Also note that the effective buffer strength of a buffer at a pH equal to it's pKa is ten times that of the same buffer at a pH titrated to a pH 1 pH unit away from it's pKa, etc. Buffer strength is usually not critical in HPLC, but it can be critical, so don't neglect it.)
I've probably neglected to mention something important here, but I'll be happy to field questions. If I don't respond quickly, it's because I don't monitor this site often.
(If indeed this is the first time this information has been "published", I expect that might be because chemists are not generally taught pH calculations, at least not for multi-pKa systems. I learned to do such calculations in my studies of biochemistry.)
Anyone who has done buffer calculations is probably aware of the "sigmoid" curve titration plot (volume reagent added versus pH). I don't believe it is commonly appreciated that this plot need not be sigmoid, but can be linear. That is to say, within a specified region of pH, it is possible to have a linear relationship between titrant and pH.
More to the point with respect to HPLC mobile phases, it is possible to have two stocks of buffer reagents, and linear change from one to the other will, within a wide range, give a linear and predictable buffer pH. (Note always that when buffer is combined with organic solvent, the apparent pH as measured by the glass electrode will shift. I am speaking of the pH before addition of any organic solvent.)
The "magic" involves in choosing a series of pKa's that are approximately 1.5 pH units apart. (I have no intention of proving this here. I have and can prove it mathematically, but more importantly, I have demonstrated it experimentally. What is important is that this is a calculated result that has been confirmed experimentally, and those calculations can be confirmed by anyone with a slide rule or computer, and those experimental confirmations can be reproduced by anyone who has an analytical balance and a pH meter.)
Curiously, citric acid is a perfect polyacid for this use, as its three pKa's differ by almost exactly 1.5 pH units. (Note that some published values for the pKa's of citric acid include one wrong value -- I forget which. I discovered this when I couldn't confirm the published pKa's, and later found other published values that agreed with my calculations.)
Hence, if you make, for example, 100 mM trisodium citrate solution, and 100 mM citric acid solution, you will find that across the midrange of possible mixtures, the pH will vary linearly with the volume fraction of the former in the mixture (or inverse linearly with the volume of the latter). At the extremes, the linearity is lost, but this still gives a wide range of pH control BY VOLUMETRIC MEANS.
(Note that citric acid can be corrosive to stainless steel and other metals. If using citric acid in HPLC systems, purge after use to prevent the mobile phase from attacking the metal. This is also true of other chelating agents and of chloride ion.)
There is no reason that a single polyacid is needed. Any series of acids or bases (so long as they don't react, precipitate, etc.) can be employed for this purpose. So, for example, one could choose three bases with pKa's of 7.5, 9, & 10.5 and create a two-component buffer system that would range linearly from about (I'm guessing here) pH 6.5 to pH 11.5. You can calibrate this if you wish, so you know the relationship between volume and pH, but it isn't necessary to do so, as it's simpler just to measure the pH of the buffer system that works best for you. This is to say that the importance of linear variation in pH is not that it's predictable, but that it is gradual and reproducible, so you can come back later and determine the exact pH you found to work best on the basis of the volume mixture you used during the chromatography.
So, how is this useful for method development? Well, ternary mixers are common. Consider reversed-phase HPLC (e.g., C18 column). Devote A to methanol, B to acetonitrile, C to 100 mM citric acid and D to 100 mM trisodium citrate. Start with some level of A, with the balance 50:50 C & D, and vary the level of A across a few chromatograms. Next use something less of B (as acetontrile is a "stronger" solvent than methanol), with the balance 50:50 C & D, and vary the level of B across a few chromatograms. A few such tests will tell what the organic can do for you, and might suggest whether a mixture of methanol and acetonitrile will be beneficial (possible, but rare in my experience). Now start with some results that look promising and vary the C-to-D ratio, while leaving the organic alone. This alters pH only, and in a linear way. If pH matters in your separation (which it may if you're dealing with acids and bases, but probably won't if you're not), then you'll see changes of interest.
Find some "optimal" results, but don't spend a lot of time on them because the next thing you'll do is to change the buffer system entirely. (As mentioned, citric acid is a poor choice as buffer due to corrosion effects. Or, if you're using a mixture of other acids or bases, you might not need all these in the final mobile phase.) Determine the ratio of C to D you prefer, pump this w/o organic until you've collected maybe 50 mL, then take it to a pH meter and read the actual pH.
Once you know the pH you need, choose a buffer that will provide this pH -- namely, one that has a pKa within one pH unit of the desired pH. (Also note that the effective buffer strength of a buffer at a pH equal to it's pKa is ten times that of the same buffer at a pH titrated to a pH 1 pH unit away from it's pKa, etc. Buffer strength is usually not critical in HPLC, but it can be critical, so don't neglect it.)
I've probably neglected to mention something important here, but I'll be happy to field questions. If I don't respond quickly, it's because I don't monitor this site often.
(If indeed this is the first time this information has been "published", I expect that might be because chemists are not generally taught pH calculations, at least not for multi-pKa systems. I learned to do such calculations in my studies of biochemistry.)