Is it possible for non-retained compounds to elute before V0

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Can analytes be so polar that the elute before the void time? I'm asking because I tested some material a while back where it looks like this happened.
Depends on exactly how you define the void volume of the column. If you take it to be the total volume of liquid, both between the particles and inside *all* of the pores, then it is quite likely that some compounds will elute before t0 due to exclusion from the pores. This can be either size exclusion (too big to get in) or ion exclusion (same charge on the analyte and the stationary phase surface).

If you go back and look at the archives (the "old" version of the Forum from 1999 - 2004), you'll find quite a discussion on how to best get an accurate measurement of t0. If I recall, D2O was suggested as a likely candidate to totally permeate all the pores (and be detected by RI).
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Appreciate the detailed respone.
3 posts Page 1 of 1

Who is online

In total there are 24 users online :: 2 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider], Semrush [Bot] and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry