Methane detection threshold

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Hi

Does anyone know the approximate detection threshold for methane with a GC with FID detector?

I have been detecting 1000ppm of methane but that seems to be the limit to what I can achieve. I was hoping to lower this although I don't know if this is realistic as I don't have much experience. Im using a GC with FID detector and a CTR dual column.

Ive tried playing with the injection/detector /column temp and pressure and Im also playing with split vs split less injection but are there any tips on how to lower the detection threshold?

Many thanks

Matt
With a packed column and 1 ml injection volume in the splitless mode, you can easily get to 1 ppm RL for methane with GC/FID.
Using a BP 1 (methyl silicone) capillary column 30m X 0.32 mm. Head pressure 15 psi, Injection temp 225 C. Column Temp 80 C, Detector Temp 250 C, Hydrogen as a carrier, split/splitless with a 2cc sample loop, I can detect .1 ppm methane easily . If you need further specs feel free to ask. Using an HP 5890 Series II.
3 posts Page 1 of 1

Who is online

In total there are 11 users online :: 0 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 11 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry