What injection volume?
What is the concentration of the main peak?
What is your sample dissolved in?
What is your actual LOQ?
How much baseline noise do you have?
What I'm driving at with the first three questions is that detectability is more closely related to mass-on-column than it is to concentration. If your diluent is weak and you don't have a huge main peak to bother with, simply increasing the injection volume can improve your sensitivity (in terms of concentration; your mass sensitivity remains the same).
The last two questions get at the signal/noise issue. LOD and LOQ are fundamentally limited by S/N. You can have very good detectability with a very small peak IF your baseline noise is low. The manufacturer's performance spec for your detector should give you an idea of how much noise to expect (take it with caution, however; the spec is usually for a longer wavelength and under no-flow conditions); you should be within a factor of 5-10 of that value. If the noise is worse than that, you may have a lamp problem, a dirty cell, erratic flow, etc.