We are representing a (P.S.A. Carbon Molecular Sieve based) "industrial" Nitrogen gas generators manufacturing company. We've recently commissioned an N2 generator designed to deliver N2 gas with 10 PPM Oxygen contents. If someone is aware of Pressure Swing Adsorption, then the N2 gas being produced (by this technique) is a mixture of itself, O2, Argon and may be some other traces in it (we don't know as our main concern is to monitor the concentration of Oxygen). Furthermore, our generator has a built-in (online) Oxygen analyzer in it.
Our customer used to buy N2 gas cylinders in the past and the gas company claims it to be 99.999% Nitrogen with a couple of O2 PPM. The cylinder gas is very pure indeed as it's being produced from liquid Nitrogen!! The problem creates when the customer collects a sample from gas cylinder and run a test on their Shimadzu GC (which they purchased a few months ago). Their GC detects the Oxygen contents as 68 PPM and if we allow the same cylinder gas to run through our integral analyzer, it also reads it around 66 PPM. It means our analyzer is working fine? On the other side, when our generator is in running / production mode, the same integral (built-in) O2 analyzer shows the O2 contents as 6 PPM. The customer then collects the gas sample from our generator in the same container and under the same ambient conditions, but their GC always detects between 1.5% - 1.7% Oxygen contents (which should actually be around 10 PPM)??
As I have already stated, it's a brand new system and we have verified that our O2 sensor + analyzer are calibrated and sensing low / high contents of Oxygen in different N2 gas streams.
I am not an expert in GC's at all ..... Is there a possibility of "wrong" method being applied on their GC (column), which is not able to detect / separate Oxygen from other traces and hence depicting a misleading composition?
Will really appreciate to have some valid arguments / opinions.