PS.mobile phase is hexane:ethanol=93:7 with several drops of TBAH
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
How long does your column last? TBAH is a strong base, even in non aqueous environment. Don't you slowly convert your column to a bare silica one? Does the solvent peak follows a logic, e.g becomes bigger with time, number of injections or is it random?
Do you know the solubility of your diol in water? If it's not too soluble maybe you can extract TBAH with water.
Or adding a SPE step to your sample preparation (using an ion exchange cartridge to trap TBAH)?
Good luck.
bhuvfe
I meant that if the column you are using is silica based (zorbax-NH2 or similar) you might have a short lifetime for such a column.What do you mean of 'a bare silica one'? How should I convert the column? Appreciate if you can give some advice.