sahram » Wed Feb 23, 2005 7:15 am
Thank you very much for your answers.
Now I can give some more information.
The pH of buffer A (50mM Ammonia plus 20mM Ammoniumbicarbonate) is around 9 or so. I check the exact value today.
The reason for the components are: Ammonia gets better ionisation and Ammoniumbicarbonate gets the right molcular ion. Ammoniumbicarbonate alone splits my peaks in two and Ammonia alone make peaks broader. So this is really a compromise. I think it can be optimized to be better.
My buffer A is always freshlyl made (same day than run) and prefiltered. The system also has a degasser, so I shouldn't have any problems with it.
My sample is clean and solved in pure water. But this cleaness/dirtyness started to worry me, because in the future I should analyze by samples from dirty matrises. I hope to find good method to clean the samples.
Last night I rised the temperature from 25 to 50 degrees and reproducibility was better but the trend is still longer and longer retention after run after run. The intresting thing was that after ten samples I ran three blanco runs (no sample, only gradient) and after that the retention time came back (or actually, little less than before) and then started to rise again. This is really tireing.
I wonder if injection volume can be one factor to that how much is the shift between two single runs.
My flow is 0,5ml/min and injection volume is 1µl or 5µl. I would prefer 5µl but now 1µl has given better results.