Is it important to take into account the density of sample and standard if both the sample and standard matrix are different.
Are we actually injecting different amount by injecting same volume for both the standard and sample?

Thank you
This depends a bit on what you are doing and what question you are trying to answer. If you have diluted yor samples and standards in a specific volume of solvent, the density issue is taken care of.
As long as all calibrations and calculations are carried out on the basis of either mass or volume you should be OK. If you mix mass and volume you can get significant bias:

Brown, R.J.C., Edwards, P.R. (2006) The effect of the physical matrix on accurate measurements using fixed volume analytical techniques. J. Sep. Sci. 29: p. 2072 - 2077.

Peter
Peter Apps
Punt,

Gas or liquid?

Best regards,

AICMM
Hi,

I am refering to liquid sample.
My calibration standard density is around 0.9 and is prepared in mg/L
Due to the low concentration , sample is injected neat without any dilution. Sample density is around 0.8.
So is it essential to correct against any density difference.

Thank you
If you have calibrated in terms of mg/L, your result would be in terms of mg/L. You are using a contant injection volume, keeping the volume analyzed constant. If we assume consistant injection volume, 100% transfer of material to the GC column and no differences in detection - you have good numbers. And if you need weight/weight results, you just use the density of the analyzed solution to calculate that.

However...

With a density of 0.8 for one and 0.9 for the other, there are a few questions to be asked. This appears to be a significantly different solvent/matrix combination between standards and samples. This suggests differences in expansion volumes in the inlet and possible differences in the results of analyte sticking to active sites in the inlet (because of differing solvent activities and area covered).

Without knowing your analyte or expected concentration - keep in mind that there are effects where matrix in samples results in active sites being masked and responses are higher for analytes in samples than in standards. (Thus people who do analysis for things like pesticides do calibrations in matrix.) And, the other possibility to consider is that other compunds in the extract may result in activity in the inlet that would cause losses of your analyte.

Differences in density suggest differences in viscosity - which can result in differences in material drawn up into the syringe and delivered to the inlet. Perhaps not much of a difference, but if we are using the liquid volume as part of the quantification, it adds to the sources of bias and variability.

With potential differences in expansion volume, the inlet purge can result in different discrimination of the analyte, depending on the volume of expansion and the purge off time.

Thus, it is important to make a standard additions curve with at least one sample to be sure that you can read back the amount of analyte added plus background.
6 posts Page 1 of 1

Who is online

In total there are 12 users online :: 0 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 12 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry