Problem with HPLC - one line has no pressure

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Hi,

After changing one of the reservoirs for my Alliance 2695, I discovered that the system won't prime (either wet or dry) on the line for which I just changed the bottle. Talked to tech support who said that if there's no pressure on any of the lines, it's a check-valve problem. But however, the system will prime on all other lines. I am kind of new to troubleshooting, but imagine that a clear place to start would be to check a filter or input line.

Any advice? Thanks for your time,

Dave

for the mobile phase filter, simply try to prime without it.

this might be too far ahead, but we once had a blown membrane in the gradient proportioning valve, GPV. If you have a suitable service contract just call the engineers out to look at it, if not, bear this component in mind.
2 posts Page 1 of 1

Who is online

In total there are 23 users online :: 2 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot], Baidu [Spider] and 21 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry