Carryover in Injections - Possible Causes?

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
Hello All,

I am running a reversed phase HPLC procedure for a drug in animal feed using 85% CH3CN:15% 125 mM aqueous acetate buffer as the mobile phase, with a typical C18 column and fluorescence detection. The sample extract is methanol acidified with 0.5% HCl. I note what appear to be significant carrover peaks in blank feeds at a residue testing level. I do not see such peaks when injecting the extractant or the mobile phase. Is this likely to come from the autosampler (Waters WISP 712) or is something happening on the column? I would appreciate advice on how to identify the source and suggestions to eliminate the problem.

Thanks very much,
CF at NDA

Have you eliminated the possibility that your blank feed may be contaminated with your drug? Have you ever gotten a blank run from your blank feed?

In the past I did not see a response at the retention time of interest, but I have not really done extensive testing. Five other non-medicated feeds from different manufacturers also showed a peak at that retention time during a recent sample run.

Thanks,
CF at NDA

You can test your autosampler in the following way: First inject a high-level standard followed by a solvent blank. If there is carryover originating in the autosampler, it will show in the blank. The WISP injector can give long service, but when the seals go out, it can also give some very strange behavior.

It is also possible that there is some extraneous co-eluting interference that your sample workup is not removing or your column is not resolving. One way to test for that is to use an alternate wavelength and see if you get the same result. A wavelength-scanning fluorescence detector is nice if you can get one.
Mark Tracy
Senior Chemist
Dionex Corp.
4 posts Page 1 of 1

Who is online

In total there are 18 users online :: 4 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Bing [Bot], Google Feedfetcher, Semrush [Bot] and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry