I am trying to figure out if running pyrolysis gasoline on a lowox column will cause any damage. I am running it on an agilent 7890A with a capilary column attached to a boiling column. I am kind of new at working on this type of thing, so any help is greatly appreciated.
It probably won't "hurt" anything. I would be cautious about co-elutions of the heavy aromatics/unsaturates with the oxygenates. If the boiling point column is set up to work as a "stripper" column, allowing backflush of the heavies after the oxygenates have eluted to the Lowox column, that'll help.
Lowox is a column based on adsorption chromatography. It has very low kapacity (loadability)meaning it will overload very fast. It works nicely for oxygenates in a volatile hydrocarbon matrix, but if higher hydrocarbons (and aromatics)are present, large amounts may impact the separation.

thats why often a precolumn (PDMS) is used to make sure the higher boilers are switched back

the column by itself can take these matrices as long as all peaks are eluting.
3 posts Page 1 of 1

Who is online

In total there are 17 users online :: 3 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Bing [Bot], Semrush [Bot] and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry