Terrible, Lingering, Un-removable GHOST PEAKS!!!
Posted: Fri Jan 07, 2011 1:58 am
PLEASE HELP!!!
I am an experienced chromatographer and have been using a new 7890 for the past 8 months. We run several high temperature analysis (semi-volatiles, triglycerides, etc) using the split inlet and FID.
Recently I have been invaded by ghost peaks that I CANNOT rid the system of. I have tried all the traditional fixes but nothing has worked so far. Details:
-The problem is not related to the ALS/syringe, as the ghost peaks appear simply when cycling the GC up to 320C with no injection.
-The ghost peaks appear on several different columns, and after trimming both ends of the column and/or reversing it.
-The FID has been cleaned and the jet swapped for a new one, the ghosts reappear.
-The liner, septum, o-ring, gold seal have all been changed to new pieces....and the ghosts return.
-The inlet shell weldment has been cleaned using a gun brush and solvents (CH2Cl2, acetone, MeOH). No improvement.
-The split vent trap has been changed. The copper tubing leading from the inlet shell weldment to the split vent trap has been changed. The ghosts are still hanging around.
-High purity solvents have been used, and backflash calculations show that I am well within the volume limits of my liner.
I am charged to figure this out as my small company is averse to costly service visits. I am at a total loss and turn the GC community for any offer of wise advice/experience. I am at the point where I am thinking about changing out the entire injector body shell, split vent assembly, and/or inlet assembly...but this is also costly and in the realm of service engineer expertise.
PLEASE HELP!!!!!!
Any advice would be much appreciated. Thanks!
PLEASE HELP!!!!!!