Effects of moisture on a GC-FID

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
Hello! This is my first post on the Chromatography Forum!

I have been running a 1 in 40 Propylene Glycol deicer to Acetone on our GC-FID and getting a strange flat topped plateau. It was recently discovered that the PG is 8% water. I was wondering if our inability to resolve the peak was due to the moisture? We are running a 30m Rtx-1 fused silica column with an id of 0.32 mm.

Thank you.
Please supply which GC you are using and the conditions, especially for the FID and it's electronics.

Gasman
RedMello06,

Neat or split? Analog out or digital out?

Best regards,

AICMM
Do you get this same flat topped peak when you inject acetone with no propylene glycol in it? (rinse the syringe carefully to avoid carryover.) If so you are looking at an overloaded solvent peak. And please post pictures of the GC traces.
4 posts Page 1 of 1

Who is online

In total there are 16 users online :: 2 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Majestic-12 [Bot] and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry