I just met this situation: the peaks are more noisy at the top, the baseline seems OK (see picture bellow).
That are hydrocarbon peaks (FID) in a Toga analysis. The GC is new.
Thanks for any inputs.
Advertisement
Discussions about GC and other "gas phase" separation techniques.
Thanks for your answer. The GC is an Agilent 7890 and brand new. Anyway, I will check the flow rates of H2, Air and see.Almost looks like the time about 30 years ago that I forgot to turn the air flow on to the FID detector. Check the basics.
Thanks AICMM for the input.thohry,
How much are you shooting? Is this a standard or is this a standard in oil or is this just the oil? You are a little noisy for an FID but not that noisy, I am guessing that you are running +/- about 0.3 pA? This is high for an FID but not outrageous. If you think this is too high, I would look at your gas supplies for contamination and then I would look for leaks.
Best regards,
AICMM
The installation test is OK, it's C14, 15 and C16 peaks and they all are OK.If the GC is new it should have some kind of guarantee on it, in which case this is Agilent's problem. Did it pass it's installation tests ?
Peter