FID signal - noise at the apex of the peak

Discussions about GC and other "gas phase" separation techniques.

8 posts Page 1 of 1
Hi all,

I just met this situation: the peaks are more noisy at the top, the baseline seems OK (see picture bellow).
That are hydrocarbon peaks (FID) in a Toga analysis. The GC is new.
Thanks for any inputs.

Image
Almost looks like the time about 30 years ago that I forgot to turn the air flow on to the FID detector. Check the basics.
Almost looks like the time about 30 years ago that I forgot to turn the air flow on to the FID detector. Check the basics.
Thanks for your answer. The GC is an Agilent 7890 and brand new. Anyway, I will check the flow rates of H2, Air and see.
thohry,

How much are you shooting? Is this a standard or is this a standard in oil or is this just the oil? You are a little noisy for an FID but not that noisy, I am guessing that you are running +/- about 0.3 pA? This is high for an FID but not outrageous. If you think this is too high, I would look at your gas supplies for contamination and then I would look for leaks.

Best regards,

AICMM
thohry,

How much are you shooting? Is this a standard or is this a standard in oil or is this just the oil? You are a little noisy for an FID but not that noisy, I am guessing that you are running +/- about 0.3 pA? This is high for an FID but not outrageous. If you think this is too high, I would look at your gas supplies for contamination and then I would look for leaks.

Best regards,

AICMM
Thanks AICMM for the input.
I injected standards, not oil and it's about 100 ppm (at max). The noise at the peak apex is about 1.5pA for the later peaks. I have put filters in the carrier gas line and it is better, just a bit.

Regards
If the GC is new it should have some kind of guarantee on it, in which case this is Agilent's problem. Did it pass it's installation tests ?

Peter
Peter Apps
If the GC is new it should have some kind of guarantee on it, in which case this is Agilent's problem. Did it pass it's installation tests ?

Peter
The installation test is OK, it's C14, 15 and C16 peaks and they all are OK.
This one has FID, Ni converter and TCD. I am thinking about the argon cylinder but not so sure.
Thanks and regards
Thanks all for replying.
It's now OK after conditioning the column, inlet etc.. and gas filtering.
8 posts Page 1 of 1

Who is online

In total there are 13 users online :: 2 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 11 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry