Alright, so I am using an Agilent 7890A GC equipped with an FPD to run a standard 25ppm sample of hydrogen sulfide in helium. The column I'm using is a GS Gas-pro 30m by .32mm column by J&W Scientific. The carrier gas is helium flowing at 8ml/min, though I've changed that rate several times and it hasn't helped me. My system has been checked thoroughly for leaks, so that isn't the problem. Let me know if you need any more information.
I've been running samples of the H2S in He in an attempt to calibrate the machine, but I haven't even been able to develop a method that gives me good data. I've been using several different methods to measure the concentration, but no matter what I do to the method, the peaks of my runs do not remain constant as I run identical samples with identical methods, instead, they continuously increase.
They (their area, but also height) increase at a decreasing rate, and I think that they may level out after what ends up being pretty much an entire day of doing runs (at least 15, probably closer to 30), though usually I get fed up and try changing the method before that. As you might imagine, running identical tests for an entire day is, well, undesirable. Typically the sample gas will be flowing around 10sccm, though I've tried several different flow rates, all with the same results.
Here's a realistic but fabricated example of the data I am getting.
Run 1: 2000pAs
Run 2: 4000pAs
Run 3: 5500pAs
Run 4: 6750pAs
... after maybe 15 more runs it might (though not necessarily will) stabilize around
Run 19: 1.8E4 pAs
Run 20: 1.79E4pAs
etc.
Typically, my method involves a ramp from a low, say 35C, up to a high temperature, say 200 or 250 C, where it remains for a few minutes in an effort to somewhat bake out the column. Lately I have been trying doing a ramp from 50 to 80C at about 6 degrees/min, with the same results.. I've tried changing the load and inject time, reducing it from .5min to .2min, with strange results. I observed two peaks, right next to each other. I'm unsure if that is relevant to my problem, though I'm really at my wit's end, so I figure I should give as much information as possible.
Initially, I thought that the increasing peaks might be due to sulfur compounds (or really, anything) remaining in the column between runs, and being picked up in the next run. To test this theory, I ran pure helium through the system, using the same method. No peak, so I feel that that's essentially ruled out.
The methods I've used are from literature on the subject, literature that doesn't mention anything along the lines of the problems I've been facing.
Really, I don't care if you have a bizarre suggestion, please, anything at this point is appreciated. If it's my machine, then fine, it can be fixed, if its a foolish error on my part, then that too can be fixed. I am going insane here, and nothing I have read has lead me in the right direction at all. Again, I'll answer any questions about my setup that you have.