Agilent 1200 LC/ 6410 MSMSMS mainpass pressure drop

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
I am running an agilent 1200 and the other night the solvent ran to dryness. Since then I am not able to achieve and maintain proper pressure with the injection valve at the mainpass position. When in the bypass position I have the proper pressure so it seems as though there is something going on in the flow path from the injection valve to the needle seat. I have flushed all lines and get get the proper flow rate thorugh this path, just not the proper pressure. Also primed the pump after dryness. Any thoughts?

Doesn't sound like a pump problem from what you describe. Is there any leakage around the needle seat in the mainpass setting, like using a tissue to see any?
Sorry for the very delayed response in posting the solution. It was in fact a leak at the needle seat. replaced this part and back to normal operation.
3 posts Page 1 of 1

Who is online

In total there are 25 users online :: 1 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider] and 24 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry