Your concern is understandable about the ability of SCX to tell the difference between a peptide with, say, 9 (+) charged residues and one with 10. That said, a chromatography column can distinguish between a single residue variation in favorable cases provided the column can see it. In this case, since your variation is near the N-terminus, then select a mode of chromatography where the N-terminus is a preferred binding site. That means SCX, per this link to a recent article on peptide orientation effects on selectivity:
http://pubs.acs.org/doi/pdf/10.1021/ac100651k
Since your peptide is so basic, then you have the option of using a WCX column, not just an SCX one, for cation-exchange. You can bind the peptides in something like 15 mM ammonium acetate, pH 5, and then run a gradient to 15% unbuffered acetic acid or else 2% unbuffered formic acid. The drop in pH will uncharge the carboxyl- groups of the WCX material, causing your peptides to elute in a convenient volatile solvent. Of course, you won't be able to monitor absorbance unless your peptide has an aromatic residue, permitting you to use longer wavelengths. Here's a link to a poster of mine from 1997 on the subject, involving our PolyCAT A columns:
http://www.polylc.com/downloads/ISPPP_1 ... poster.pdf