Separating nearly identical peptides

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

5 posts Page 1 of 1
I'm having trouble separating two peptides- both > 80 aa and 97% identical.

One has S, the other RY (near the N term), but are otherwise identical.
These peptides are very well retained on RP columns, and are have a basic pI, 2 lys + 7, 8 Arg.

I can sort of resolve these on a C4 column using HFBA + MeCN, but the peaks are extremely broad due to a very shallow gradient (0.14% MeCN/min). It is hard for me to tell if they are completely resolved or not due to this.

Since these peptides differ in number of + charges (+9, +10), would SCX be a better choice? Will I run into the same problem (broad, poorly shaped peaks) from having to use a shallow gradient?
I suppose you have tried TFA and/or different pH's (triethylamine acetate)?
A. Carl Sanchez
I have tried TFA.

I went with HFBA as ion pairing agent in hopes I could exploit the difference in Arg content. You expect increased retention for the peptide containing the extra arg, However, I suppose a neutral pH or something like phosphoric acid- could invert order of elution. Perhaps this is worth trying.

I guess my main question could be re-phrased as:

I can't make the gradient any more shallow in my C4 HFBA RP method, where the center of each peak is < 2 minutes apart, but they are so broad, I fear they over-lap. Is the charge difference between a +9 and a +10 peptide enough of a difference to be able to resolve these peaks by > a few minutes on SCX?
Your concern is understandable about the ability of SCX to tell the difference between a peptide with, say, 9 (+) charged residues and one with 10. That said, a chromatography column can distinguish between a single residue variation in favorable cases provided the column can see it. In this case, since your variation is near the N-terminus, then select a mode of chromatography where the N-terminus is a preferred binding site. That means SCX, per this link to a recent article on peptide orientation effects on selectivity: http://pubs.acs.org/doi/pdf/10.1021/ac100651k
Since your peptide is so basic, then you have the option of using a WCX column, not just an SCX one, for cation-exchange. You can bind the peptides in something like 15 mM ammonium acetate, pH 5, and then run a gradient to 15% unbuffered acetic acid or else 2% unbuffered formic acid. The drop in pH will uncharge the carboxyl- groups of the WCX material, causing your peptides to elute in a convenient volatile solvent. Of course, you won't be able to monitor absorbance unless your peptide has an aromatic residue, permitting you to use longer wavelengths. Here's a link to a poster of mine from 1997 on the subject, involving our PolyCAT A columns: http://www.polylc.com/downloads/ISPPP_1 ... poster.pdf
PolyLC Inc.
(410) 992-5400
aalpert@polylc.com
I bet the case could easily be handled by HIC.
I've used HIC for separation of extremely similar peptides/proteins (typically degradants) and it never failed me.

Best Regards
Learn Innovate and Share

Dancho Dikov
5 posts Page 1 of 1

Who is online

In total there are 367 users online :: 2 registered, 0 hidden and 365 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider], Google [Bot] and 365 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry