The USP monograph specifies the different LC analytical methods (i.e. the mobile phase and the gradient elution are different) for "Assay" and related compounds" in drug substance. Why the LC analytical method between "assay" and "related compounds" in USP monograph is diffenence, even the API peak can be detected as using the method of "related compounds" ? Could I just use the "related compounds" method of drug substance to test the " Assay" and "related compounds" of drug product if USP monograph doesnot specify the drug product ? Thank you.
They certainly don't HAVE to be different chromatographically (column, mobile phase, gradient) if everything is separated. They MAY have to be different because of sensitivity/quantitation requirements.
It is not unusual to have to inject more sample (higher conc or more volume) to quantitate related substances or impurities, so much that the API is off scale or out of the linear range.
From a time/efficiency standpoint it should be noted, as assay method does not need to separate all the related substances from each other, only from the API. This means that Assay method can be isocratic and much faster. That can be important for doing thing like content uniformity where you need to assay a lot of samples.
I've had situations where i used single method for purity and potency (assay) and others where they were separate.
- Karen