We have a V3600 that is taking FOREVER to cool. We've confirmed that trap door that opens to exhaust the hot air is not working.

Anyone have experience replacing this? Not sure what the status is with Bruker/Varian re - service and spare parts.

(edit) Ran through the self test on the GC. It says that the temperature control board isn't supplying the proper voltage for the cooling solenoid to function. Seems that we have three spare temp control boards and all three have a note affixed "won't control cooling solenoid". Tested all three to no avail. :(
Hi,

You can contact your local Bruker Chemical Analysis office for information about spare parts etc. But they can provide you almost all the part Varian did in the past.

Good luck.