I try to develop a method for amino acids. The separation is fine, but the lifetime of the column is dramatic.
I use a Luna column and I run a gradient from 4%B to 75% B. Mobile phase A is 20 mmol phosphate pH 7.0 and mobile phase B is ACN. Column temperature is 40 degrees.
I inject 1uL sample, where the sample is dissolved in 100mmol sodium borate pH 9.0.
What happens you can see in the figure below. There is the performance shown of a column test mixture before using it for the amino acids and after approximately 100 injections. The retention time remains the same, just the peaks are spitted. If I reverse the column, the performance is much better again. The pressure was in both cases equal.

At this time I have several ideas about the nature of the problem:
Blocked inlet frit ( but I would expect that the pressure increases).
Voids due to mechanical shocks (which I didn’t saw in the pressure profile).
Silica dissolved random inside the column due to the high pH of the mobile phase.
Silica dissolved at the top of the column due to the high pH of the injection solvent ( I opened the column but wasn’t silica missing at the top).
Precipitation by phosphate/ACN contact
Too long equilibration at low %B with phosphate buffer..
I’m very interested in your opinion what would be the most vital explanation (and which above causes would not be likely).
Of course I also look for a solution. I have a few idea’s
The use of a high pH resistant column (like Xbridge or maybe zorbax eXtend).
The use of a guard column
Use an alternative (organic buffer) instead of phosphate.
….
Again what would be the most logical next steps.
Thanks in advance
I use a Luna column and I run a gradient from 4%B to 75% B. Mobile phase A is 20 mmol phosphate pH 7.0 and mobile phase B is ACN. Column temperature is 40 degrees.
I inject 1uL sample, where the sample is dissolved in 100mmol sodium borate pH 9.0.
What happens you can see in the figure below. There is the performance shown of a column test mixture before using it for the amino acids and after approximately 100 injections. The retention time remains the same, just the peaks are spitted. If I reverse the column, the performance is much better again. The pressure was in both cases equal.

At this time I have several ideas about the nature of the problem:
Blocked inlet frit ( but I would expect that the pressure increases).
Voids due to mechanical shocks (which I didn’t saw in the pressure profile).
Silica dissolved random inside the column due to the high pH of the mobile phase.
Silica dissolved at the top of the column due to the high pH of the injection solvent ( I opened the column but wasn’t silica missing at the top).
Precipitation by phosphate/ACN contact
Too long equilibration at low %B with phosphate buffer..
I’m very interested in your opinion what would be the most vital explanation (and which above causes would not be likely).
Of course I also look for a solution. I have a few idea’s
The use of a high pH resistant column (like Xbridge or maybe zorbax eXtend).
The use of a guard column
Use an alternative (organic buffer) instead of phosphate.
….
Again what would be the most logical next steps.
Thanks in advance