H2 gas analysis with TCD, Argon carrier gas
Posted: Tue Dec 07, 2010 8:55 pm
I have submitted gas samples for GC analysis for O2, N2, CO2, CO, C1-C4 hydrocarbons and H2S to two different labs. The results of analysis between the labs (using direct sample injection from canister to instrument from duplicate samples), both using a GC FID/TCD, do not agree. One lab reports H2 gas 2-3 orders of magnitude greater than the other.
The analytical procedure for the one lab, reporting [H2] of 1700 ppm is as follows:
Sequence of analysis: CO2, O2, N2, C1-C4 were run then the sulphurs and finally the H2.
- The samples were not diluted when the CO2, O2, N2, C1-C4 were run. The canisters were pressurized just before the sulphurs were analyzed.
- GC Model: 7890A GC FID/TCD
- Oven temperature: Run starts at 40 and goes to 200 at varying ramps. H2 comes out between 140 and 200
- Single inlet
- Argon Carrier Gas
- 1 mL Sample loop volume
- Backflushing time: 0.5min with carrier gas
- Column specs: Supelco 60/80 Carboxen 1000 15’X1/8” SS (2.1mm ID)
- 2X dilution with ultra high purity nitrogen
The other lab reports [H2] of 810,000 ppm. I am waiting to get details of analytical procedure for this lab.
Can you suggest possible gas interferences, or analytical errors that might cause this 2-3 order of magnitude discrepancy in Hydrogen gas?
Thank you!
The analytical procedure for the one lab, reporting [H2] of 1700 ppm is as follows:
Sequence of analysis: CO2, O2, N2, C1-C4 were run then the sulphurs and finally the H2.
- The samples were not diluted when the CO2, O2, N2, C1-C4 were run. The canisters were pressurized just before the sulphurs were analyzed.
- GC Model: 7890A GC FID/TCD
- Oven temperature: Run starts at 40 and goes to 200 at varying ramps. H2 comes out between 140 and 200
- Single inlet
- Argon Carrier Gas
- 1 mL Sample loop volume
- Backflushing time: 0.5min with carrier gas
- Column specs: Supelco 60/80 Carboxen 1000 15’X1/8” SS (2.1mm ID)
- 2X dilution with ultra high purity nitrogen
The other lab reports [H2] of 810,000 ppm. I am waiting to get details of analytical procedure for this lab.
Can you suggest possible gas interferences, or analytical errors that might cause this 2-3 order of magnitude discrepancy in Hydrogen gas?
Thank you!