sepration of 2,4 hexadiene in benzene matrix
Discussions about GC and other "gas phase" separation techniques.
5 posts
Page 1 of 1
I want to determined 2,4 hexadine in benzene matrix. The commercially available of hexadine consist of three isomers. Two of isomers are very close to other non aromatics of benzene. I found that the best separation is in low temperature and lowest temperature possible in our GC is 28 C◦ . Suppose one isomer appear at 13.24 and nearest non aromatic appear at 13.39. Also second isomer appears at 14.157 while a big non aromatic peak appears at 14.00. These two pair of peak has different properties. In case of low column flow (1.2 ml/min and initial pressure 28 psi) first pair has best separation but second pair not good separate. By increasing column flow second pair will be separate but first pair converts to one peak. We are using Varian CP-3800 and it has an option for pressure programming. I tried to combine two above fact. That mean use a program with low pressure until end of first two pairs and increasing pressure (column flow as result) in order to separation of second peaks. But it wasn’t successful. Is any idea for better results? As recommended by basic method we are using two columns in series for this test. One CP-WAX 52 CB 60 m 0.32 followed by one CP SIL 8 CB 30 m 0.25.
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
Using a 3 column Backflush heartcut application, should separate them cleanly.
A short polar column to move BZ from HDs, backflush BZ and cyclo and heavier alkanes, a long non-polar column to move lighter alkanes ahead of HDs, sending them to vent, but cutting the HDs onto a long polar column to move olefins and medium alkanes ahead of HDs.
Adjustments of each column length or film thickness may be necessary.
Good luck,
Rod
A short polar column to move BZ from HDs, backflush BZ and cyclo and heavier alkanes, a long non-polar column to move lighter alkanes ahead of HDs, sending them to vent, but cutting the HDs onto a long polar column to move olefins and medium alkanes ahead of HDs.
Adjustments of each column length or film thickness may be necessary.
Good luck,
Rod
-
- Posts: 1773
- Joined: Fri Feb 09, 2007 10:33 pm
mr3573,
Are you literally using two columns connected to each other? If so, the separation you gain on one could easily be lost on the second (to the same compounds or to different compounds.) Much better to use a valve to make a cut from first to second (or, as Rodney suggests, a heartcut is even better.) Considering the temperatures you are working at, a switching valve would be a relatively easy thing to implement.
Best regards,
AICMM
Are you literally using two columns connected to each other? If so, the separation you gain on one could easily be lost on the second (to the same compounds or to different compounds.) Much better to use a valve to make a cut from first to second (or, as Rodney suggests, a heartcut is even better.) Considering the temperatures you are working at, a switching valve would be a relatively easy thing to implement.
Best regards,
AICMM
-
- Posts: 36
- Joined: Sat Dec 11, 2010 8:32 am
Dear chromatographer1
sorry for question but what means HDs ?
sorry for question but what means HDs ?
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
HDs = hexadiene isomers
Rod
Rod
5 posts
Page 1 of 1
Who is online
In total there are 31 users online :: 0 registered, 0 hidden and 31 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 31 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 31 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science