Unusually shaped peaks in chromatogram
Posted: Tue Dec 07, 2010 5:59 pm
Dear fellows,
I'm currently working on the validation of a GC-ECD method for the determination of THMs in drinking water. The preconcentration is through LLE and the seperation of the compounds takes place on a 60m long capillary column. The peaks have a very unusual appearance: each peak starts off normally and sharply, reaches its apex and then drops extremely slowly. Near the end of the peak there's a sharp drop to the baseline creating something that looks like a step. Do you think the slow return to the baseline is due to the length of the column? Has anyone else observed that final drop? Since the peak is extremely assymetric (almost impossible to decide where it stops) would you recommend using the peak height instead of its area as the response?
Thanks in advance for any valuable comment.
I'm currently working on the validation of a GC-ECD method for the determination of THMs in drinking water. The preconcentration is through LLE and the seperation of the compounds takes place on a 60m long capillary column. The peaks have a very unusual appearance: each peak starts off normally and sharply, reaches its apex and then drops extremely slowly. Near the end of the peak there's a sharp drop to the baseline creating something that looks like a step. Do you think the slow return to the baseline is due to the length of the column? Has anyone else observed that final drop? Since the peak is extremely assymetric (almost impossible to decide where it stops) would you recommend using the peak height instead of its area as the response?
Thanks in advance for any valuable comment.