I try to determine some solvents (methanol, Ethanol, Ethyl Acetate, toluene and DMSO) by GC-FID. I use NMP as diluent. My problem is that the diluent gives a not stable peak at the same retention time with DMSO. I changed some parameters of GC (for example the temperature program to the oven) but the problem was not solved.
Any suggestion please?
Two of the obvious come to mind.

Use a different Diluent.

Use a different column.

Don't insist on science working as YOU want it to work to your convenience.

Sometimes, we ourselves have to realize that on occasion, sometimes:

You eat bear.

The bear eats you.

In other words, you may require a second method to do DMSO analysis using methyl alcohol, or methyl acetate or methylene chloride or some other solvent as a diluent.

best wishes,

Rod
Hi

Well some litterature/catalouge searching would perhaps avoided the problem.

Column manufactures like Agilent/J&W typically has retetion indexes for DB-1, WAX and 624 columns for the most common solvents.
Izaak Kolthoff: “Theory guides, experiment decides.”
3 posts Page 1 of 1

Who is online

In total there are 59 users online :: 1 registered, 0 hidden and 58 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Baidu [Spider] and 58 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry