Not having much luck with sample prep for glutaraldehyde...can anyone help me find this article instead?
Determination of Formaldehyde and Other Aldehydes in Industrial Surfactants by Liquid Chromatographic Separation of Their Respective 2,4-Dinitrophenyl-hydrazone Derivatives
James R. Dahlgren and Melvin N. Jameson, “Determination of Formaldehyde and Other Aldehydes in Industrial Surfactants by Liquid Chromatographic Separation of Their Respective 2,4-Dinitrophenyl-hydrazone Derivatives,” Journal of the Association of Official Analytical Chemists (Vol. 71, No. 3, 1988), pp. 560-563. I have this at work. We also use the following for glutaraldehyde.
DNPH reagent (0.12 % 2,4-dinitrophenylhydrazine in HPLC acetonitrile): weigh 0.11 – 0.13g 2,4-dinitrophenylhydrazine into a 100 ml volumetric flask; mix and dilute to volume with acetonitrile (mix well to dissolve completely).
Formaldehyde hydrazone standard (2,4-dinitrophenylhydrazone): pipet 1.00 ml formaldehyde standard solution above into a 20 ml scintillation vial; then add 4.0 ml DNPH reagent and 100 ul of 2 N HCl. Cap and mix well; allow 10 minutes derivatization time. Filter a portion through a 0.45u disposable HPLC sample filter unit into a sample vial and cap.
Make fresh daily.
Reagent blank: pipet 1.00 HPLC H2O into a 20 ml scintillation vial; then add 4.0 ml DNPH reagent (using an Eppendorf pipet or repeating dispenser) and 100 ul of 2 N HCl. Cap and mix well; allow
10 minutes derivatization time. Filter a portion through a 0.45u disposable HPLC sample filter unit into a sample vial and cap. Make fresh daily.
Weigh 1 g well-mixed liquid sample into a 20 ml scintillation vial. Add 4.0 ml DNPH reagent and 100 ul of 2 N HCl. Cap and mix well; allow 10 minutes derivatization time. Filter a portion through a 0.45u disposable HPLC sample filter unit into a sample vial and cap.
If the contribution of the formaldehyde hydrazone peak in the reagent blank is significant, subtract its average area from both the average Areastandard and the average Area sample before doing the calculations. If the size of the formaldehyde hydrazone peak in the reagent blank is so large that it would negatively affect quantitation of the formaldehyde level in the sample, purify the 2,4-dinitrophenylhydrazine by recrystallizing as specified in EPA method #8315.