I was wondering if anyone has had a similar problem as mine;
The samples (not diluted with MF) and the resolution solution (diluted with MF) have 2 peaks (one from API and the other from impurity). The API peak has RT=15min and the Imp peak has RT=12min. But, when I run the individual evaluation standards 0,2% API and 1% Imp (diluted with MF) their RT's are 4 minutes late?!
MF or any other of whatsoever change was not been made while analysis was running. Same MF was used for the system and for diluting the solution. The method is isocratic,, flow 1ml/min! Column T is ambient, and the samples are not being cooled. Inj.volume 20mcl. Column Superspher Select B 250x4.
The analysis was also conducted on a different system and the results were the same: RT of peaks in the individual evaluation standards are 4 minutes late than those in the samples...
Any ideas?

Tnx
Natasha
What are the compositions of the mobile phase and sample diluent? Is there a pH effect going on, ion pairing, something like that? Have you tried dissolving your samples in MP?
Where can I buy the kit they use in CSI?
The composition is:
MF A: Octan-1-sulphonic acid Sodium salt and glacial acetic acid in water (70%)
MF B: ACN (30%)

By the way this is a method that it is being used for a couple of years now (it is pharmacopoeal method), and till now there were no problems of whatsoever...

But I have noticed that the ion-pairing agent that we used for this analysis it's a brand new container and a different lot.no than the ones used before...
Natasha
But I have noticed that the ion-pairing agent that we used for this analysis it's a brand new container and a different lot.no than the ones used before...
That reagent, and preparing both in matching solvents, would be where I'd start my investigation.
Yep, also classic problem solving techniques - remake the MPs and samples
Where can I buy the kit they use in CSI?
I suggest that you mix two solutions (spike) to make sure that you compare apples to apples.
6 posts Page 1 of 1

Who is online

In total there are 8 users online :: 0 registered, 0 hidden and 8 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 8 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry