Two very similar substances: need column advice
Posted: Mon Nov 29, 2010 3:15 pm
by LCFlo
Dear all,
I have two substances, that are very similar: Estriol and 9,11-Didehydroestriol (see picture).
What type/method (RP-mode only) would you recommend for the separation of these two substances?
Thank you very much for any hint.
Regards
Florian
Posted: Mon Nov 29, 2010 5:19 pm
by Consumer Products Guy
Not exactly what you asked, but personally I'd try capillary GC after trimethylsilyl derivatization.
Posted: Tue Nov 30, 2010 12:35 am
by tom jupille
The extra hydrogen on 9 is obviously a typo. :wink:
If those are the only two compounds to be separated, then a phenyl column might give a bit better selectivity than a C18 for the presence/absence of that double bond. If there are other compounds in the sample as well, then it's a toss-up.
Posted: Tue Nov 30, 2010 8:43 am
by Alex Buske
separating Estriol and 9,11-Didehydroestriol by HPLC isn't very complicated. Most C18-Columns (150mm, 3µm) will do. Some will also separate 9,11-Didehydroestriol and 6-Didehydroestriol.
Alex
Posted: Tue Nov 30, 2010 8:56 am
by LCFlo
Hello,¨
DawnRazor: you are right: the H-atom at position 9 is wrong as noticed by Tom Jupille.
Alex: your statement sounds as you might be experienced in this kind of separation. Do you have an application or similar?
Thank you very much to all responders.
Regards
Florian
Re: Two very similar substances: need column advice
Posted: Tue Nov 30, 2010 4:03 pm
by cody84
Kinetex PFP! I just was at the presentation for Kinetex columns.
Posted: Tue Nov 30, 2010 5:45 pm
by unmgvar
I would try two possibilities
both from cosmosil for exactly those types of molecules
I would start by the Cholester type. I know of an application for vitamin D2 and D3 where the column helps splendid and the separation is super simple with MeOH Water
another possibility is Pi-NAP to try and do separation on the pi-pi in the molecules but this column uses naphtylethyl group which gives better solution then a phenyl column
the advantages of these columns in these applications is the use of simpler mobile phases with less modifiers
Posted: Thu Dec 02, 2010 6:36 am
by XL
A related separation can be found in the 2nd chromatogram in the following link: http://www.dionex.com/en-us/products/co ... 87227.html. The column is Acclaim Phenyl-1, 3-um column.
Posted: Thu Dec 02, 2010 1:05 pm
by Alex Buske
Florian,
I don't have a recent method on Estriol. If I had, it would be IP of my employer anyway. I would go with a C18 or C8 and try methanol and ACN as organic additives. Small pore size materials can have advantages.
Alex