Advertisement

Baseline fit algorithm

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

3 posts Page 1 of 1
I am trying to develop a baseline fit algorithm since last august. All attempts have failed. My baseline is simply going over the signal. I use spline fit for constructing the baseline.

Can anyone suggest a better method?

If you have insufficient information, you have insufficient information and any resulting quantitation is little better than guesswork. The only way to improve the situation is to improve the chromatography or the sample cleanup. :cry:

I'm not trying to be sarcastic here. You can't compute your way around inadequate chromatography, so you may as well do simple linear interpolation between key points.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

Hi sana_sankar,

are you using a mass spectrometer?
If yes, you could have a look at the source code of the simple baseline detector I've developed for OpenChrom.
But seriously, it is very simple and shows some inaccuracies:

http://sourceforge.net/projects/baselinedetec

Image
Simple baseline detection using OpenChrom


Further on, I do agree with tom, that it's necessary to pre-optimize the chromatography or to do a sample cleanup.
Anyhow, it only works if you are able to measure the samples again.
In other cases, a post-optimization is inevitable. That's why I've implemented a feature to include "Filters" in OpenChrom.
The latest filter is a denoising filter, to reduce chemical noise (a paper is on its way):

Image
Baseline after applying the chemical noise reduction filter


It hopefully helps you and gives you inspiration.
There is also a paper I recommend:

SUN, X. Y.; SINGH, H.; MILLIER, B.; WARREN, C. H. & AUE, W. A.
Noise, Filters And Detection Limits
Journal Of Chromatography A, 1994, 687, 259-281




Cheers,
eselmeister
3 posts Page 1 of 1

Who is online

In total there are 31 users online :: 0 registered, 0 hidden and 31 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 31 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry