Internal standard and injection volume, sensivity of the det
Posted: Sun Nov 28, 2010 10:39 am
It is common to use the same injection volume for calibration and test samples (unknown). But if you use internal standard, it should be possible to use different injection volumes, because Area ratio (substance to analyze/internal standard) is independent from injection volume.
In my case, the peak of the highest calibration point is very large (out of range, cap), so it could be convenient to inject a smaller volume for this calibration level. For the lower calibration points, the higher volume is better to achieve optimal sensitivity.
On the other hand, you could reduce the sensivity of the detector (adjust smaller gain) and inject always the same volume.
What is the best procedure?
In my case, the peak of the highest calibration point is very large (out of range, cap), so it could be convenient to inject a smaller volume for this calibration level. For the lower calibration points, the higher volume is better to achieve optimal sensitivity.
On the other hand, you could reduce the sensivity of the detector (adjust smaller gain) and inject always the same volume.
What is the best procedure?