Internal standard and injection volume, sensivity of the det

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
It is common to use the same injection volume for calibration and test samples (unknown). But if you use internal standard, it should be possible to use different injection volumes, because Area ratio (substance to analyze/internal standard) is independent from injection volume.


In my case, the peak of the highest calibration point is very large (out of range, cap), so it could be convenient to inject a smaller volume for this calibration level. For the lower calibration points, the higher volume is better to achieve optimal sensitivity.


On the other hand, you could reduce the sensivity of the detector (adjust smaller gain) and inject always the same volume.

What is the best procedure?
In my case, the peak of the highest calibration point is very large (out of range, cap), so it could be convenient to inject a smaller volume for this calibration level. For the lower calibration points, the higher volume is better to achieve optimal sensitivity.
I would not recommend this. Besides possible accuracy issues , I assume the range of the calibration curve was set where it is because samples could be that high...

So you would either need inject all samples at both volumes or re-inject off scale samples... which is far from optimal!
On the other hand, you could reduce the sensivity of the detector (adjust smaller gain) and inject always the same volume.
If possible I would recommend this.

Doing this might give you a better signal to noise ratio at the low end despite smaller peaks.

Why are you using such a board calibration curve?

In my experience, when I have needed broad calibration curves maximum accuracy was not required at the low end.

I've not done much IS work, but when I have it's been for a relatively narrow calibration range.

I have always wondered how accurate the IS method would be when using a broad calibration even if you are in the linear range of the detector... Big differences in peak areas between unknowns and IS means potentially large measurement variability because on uncertainties in BOTH the numerator and the denominator.


- Karen
2 posts Page 1 of 1

Who is online

In total there are 29 users online :: 0 registered, 0 hidden and 29 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 29 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry