Page 1 of 1

what make FILTERS paper different ?

Posted: Thu Nov 25, 2010 5:10 pm
by Ronaldo
Hi ,

could you illustrate this for me please as i don't found exactly what i need in the net . We use usually PTFE AND NYLON ( ALL 0.2 MICRON ) before injection , but why when we use No.1 or Glass fibre what factr control this and make filter differ from one to another ?

Thanks

Posted: Thu Nov 25, 2010 5:30 pm
by tom jupille
Any time you filter a sample, you run two risks:
- loss of your analytes by adsorption onto the filter
- addition of compounds leached from the filter material.

Paper is the most heterogeneous of the filter media you mentioned, and so presumably has the highest potential for adding impurities.

Paper and glass fiber are both relatively polar (paper because of its water content, and glass fiber because of surface silanols); polar analytes may stick to either (especially if your diluent is an organic solvent).

PTFE and nylon are less polar, so non-polar compounds may stick (especially if the diluent is highly aqueous).

Of course, in many cases, it doesn't matter :wink:

Posted: Thu Nov 25, 2010 9:03 pm
by JGK
Also the PTFE and Nylon membranes have a specific pore size in your case 0.2 µm

Glass fibre filter paper will probably have an effective equivalent pore size of 1 - 2 µm for liquids.

Posted: Fri Nov 26, 2010 3:09 pm
by Ronaldo
Thanks really appreaciated