You calculated RRF, let's call them response factors, the response of the analyte per response of the internal std based on an equal wt/wt ratio.
Let say for example, hexane's response is twice the response of methanol based on a weight per weight sample. RRF = 0.5 for MeOH/C6
In other words is a response for methanol at 1mg/mL is 100,000 counts then the response for hexane at 1mg/mL is 200,000 counts.
Once you know the response factor, then say in a sample of unknown sample concentration, the hexane peak area is 50,000 counts. The methanol peak (ISTD) in your dissolution solvent still gives you 100,000 counts.
That means the concentration of the hexane in the sample is 0.5mg/mL.
But you don't know the concentration of the sample that gives you a hexane concentration of 0.5mg/mL.
But if you know the concentration of the hexane in the sample on a wt/wt basis then you can calculate the wt/vol concentration of the sample solution.
Let's say you know the sample contains 2 % hexane, then the unknown sample concentration is 50x that of the hexane peak of 0.5mg/mL, and the sample solution concentration is 25 mg/mL (50 times 0.5 mg/mL)
To calculate the concentration of the sample you must know the concentration of the solvent analyte based on a wt/wt basis and you must know the response relative to a known concentration internal standard.
Otherwise, you cannot calculate the unknown concentration of the sample in solution.
best wishes,
Rod