Help!! Anyone ever seen Empower 2 inject sequence incorrect?
Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.
4 posts
Page 1 of 1
It was brought to my attention that an injection had a time and date stamp indicating that it was not injected in the proper sequence. The sample should have been injected after a bracketing standard, as indicated by the sample set but was time stamped before any of the other injections. has anyone seen this before?
-
- Posts: 580
- Joined: Fri Mar 16, 2007 11:29 pm
Never observed sth like this.
How is the rest of the sequence?
any automatic time adjustments of the PC (by network) turned on?
what about the "inject immediate" function of the sample set? (never used this, so I'm not that familiar what it does and when it is used)
How is the rest of the sequence?
any automatic time adjustments of the PC (by network) turned on?
what about the "inject immediate" function of the sample set? (never used this, so I'm not that familiar what it does and when it is used)
-
- Posts: 426
- Joined: Mon May 19, 2008 10:09 am
I don't know which setup you have (stand alone, enterprise, workgroup), but I saw this on our enterprise once, where we had some injections made in the year 1900.
This was caused by a wrong time on our server (which also synchronised all the other pc's).
After fixing this, we had some samples set which runned appearantly more than 100 years!
Ace
This was caused by a wrong time on our server (which also synchronised all the other pc's).
After fixing this, we had some samples set which runned appearantly more than 100 years!
Ace
-
- Posts: 38
- Joined: Fri Apr 09, 2010 1:52 pm
Update:
The sequence was not injected out of order. The lace went into buffering which caused the injections to get time stamped by the lace and not the server. unfortunately, the lace was set to the wrong time zone. Empower displays the time of the lace but with the wrong timezone listed for the time. We updated the time zone on the lace and all is well.
Thanks for the replies.
The sequence was not injected out of order. The lace went into buffering which caused the injections to get time stamped by the lace and not the server. unfortunately, the lace was set to the wrong time zone. Empower displays the time of the lace but with the wrong timezone listed for the time. We updated the time zone on the lace and all is well.
Thanks for the replies.
4 posts
Page 1 of 1
Who is online
In total there are 351 users online :: 0 registered, 0 hidden and 351 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 351 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 351 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science