Agilent 1100 UV-Vis VS diode array, sensitivity?

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
I've heard contrasting claims; some say the Uv-vis module is more sensitive, some say the diodearray is, can anyone shine some light at this? If the latter is true, the diode array seems like the way to go, no?

UV/VIS detectors used to be a lot more sensitive than PDAs, now they're just a (very) little bit more sensitive than PDAs. If cost is not a concern (PDAs are also a bit more expensive), and you anticipate no issues with training end users to use them properly, PDA is a more attractive way to go than they used to be.
Thanks,
DR
Image

Thanks, I was just wondering as we want to purify/analyze trace amount of material <0.3ug. So an Agilent 1100 with a DAD (I guess they have two different versions), should be the way to go.

Probably correct, though I think that, by now, you may be limited to the 1200 series.
Thanks,
DR
Image
4 posts Page 1 of 1

Who is online

In total there are 17 users online :: 2 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot], Semrush [Bot] and 15 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry