Agilent 1100 and MassLynx 4.1
Posted: Tue Nov 16, 2010 8:03 pm
by douglasgm
If I want to control an Agilent 1100 using MassLynx 4.1, do I need to connect the Agilent to the PC with GPIB cables, or can I use a jet direct card and go through cat 5? Also, If I need GPIB, do I need to daisy chain all the Agilent modules together with GPIB?
Posted: Wed Nov 17, 2010 9:20 pm
by nkg
you connect the agilent system to the pc using a gpib cable, the agilent system is connected using the can bus, from masslynx you select which device the gpib cable is connected to and set the relevant addresses for each module from the masslynx inlet editor.
Re: Agilent 1100 and MassLynx 4.1
Posted: Tue Jul 02, 2013 2:05 pm
by andy_s
Hello people,
Any idea whether this can be done with a generic (NI) GPIB card, rather than the specific Agilent one Waters seem to insist/have tested it on?
Cheers,
Andy
ps (sorry - posted this elsewhere, but wondered if people originally posting this might have a better idea- viewtopic.php?f=4&t=22929&p=109569&hili ... nx#p109569)