Electron 70eV
Discussions about GC and other "gas phase" separation techniques.
4 posts
Page 1 of 1
Guys, how can I set the acceleration energy to 70eV in 5972.
-
- Posts: 1773
- Joined: Fri Feb 09, 2007 10:33 pm
faris,
I believe it is fixed in the 72's. I don't think you had the option until the 75's but this is probably easily answered with a visit to Agilent's web site.
Best regards,
AICMM
I believe it is fixed in the 72's. I don't think you had the option until the 75's but this is probably easily answered with a visit to Agilent's web site.
Best regards,
AICMM
-
- Posts: 1680
- Joined: Sat Aug 23, 2008 12:04 am
I am not sure if you are talking about accelerating the electrons from the electron gun or the formed ions from the ion source.
As I recall, the default ionization energy (the result of acceletating electrons) is 70 eV. I have changed the voltage in a 72 or a 73 (And I am pretty sure it was a 72), but it was about 10 years ago and without access to one of those instruments now, I cann't tell you how I did it.
By the way the objective was to see if I could do soft ionization - which worked, kind of. If you drop the voltage low enough, you do get primarily the molecuar ion formed - apparently. The trouble is the signal has dropped to next to nothing...
If you want to change the accelaration of ions out of the ion source and into the analyzer, you would change the voltage on the repeller.
As I recall, the default ionization energy (the result of acceletating electrons) is 70 eV. I have changed the voltage in a 72 or a 73 (And I am pretty sure it was a 72), but it was about 10 years ago and without access to one of those instruments now, I cann't tell you how I did it.
By the way the objective was to see if I could do soft ionization - which worked, kind of. If you drop the voltage low enough, you do get primarily the molecuar ion formed - apparently. The trouble is the signal has dropped to next to nothing...
If you want to change the accelaration of ions out of the ion source and into the analyzer, you would change the voltage on the repeller.
4 posts
Page 1 of 1
Who is online
In total there are 23 users online :: 1 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 22 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 22 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science