I was wondering if anyone has problems with endrin/ddt breakdown on Agilent 6890? I have had this problem for 2 years and cannot solve it. I have changed inlet body, tried new liners, changed gas lines lowered injection temperature from 275 to 250 and the only thing that works is using new liners. And that only lasts two days maximum before I have to perform maintenence again and I am running relatively clean samples. Sometimes endrin breaks down right after running standards.
My second source standard and lowest standard read perfect.
It is too expensive to use new liners every few days.
Does any one have any other suggestions?
Thanks for your help.