Negative Peak with PFPD where H2S should be

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
We just completed calibration of our PFPD (Varian CP3800) using hydrogen sulfide (25ppm) in air calibration standard. Peak looks fine.
We then ran several natural gas field samples and all show "negative peak" at the exact retention time of H2S.
Can anyone help explain this?

I've seen this with highly concentrated samples, although it is usually a positive peak that drops to negative part way through then jumps to positive again. Obviously an overloading issue in my case.

My suggestion would be to run the sample diluted.

Rich
"Can't be king of the world
if you're slave to the grind"

NorthernGC,

I don't know the PFPD all that well but I would be suspicious of a quench (based on my discussions of FPD.) I also don't know what column so I would guess either CO2 or perhaps a large concentration of ethane as my first guess. Do you have a different column available?

Best regards,

AICMM
3 posts Page 1 of 1

Who is online

In total there are 15 users online :: 0 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 15 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry