Running a packed column on a split injection system. (I know already that I shouldn't be doing this but shimadzu has an adapter that allows for it and I'm stuck with what I have.) - split and purge vents are off.
Very simple run. Compounds are Cetyl and Stearyl Alcohol in 200 proof ethanol. 5 injections of a standard only used for resolution requirements, 2 injections of 1 sample prep. Calc is area percent.
Couldn't be happier with my standard precision, both peaks area RSD's under 1.2%.
Problem rears its head with the sample injections. The first one has noticable drop of in Area and Height. Tailing increases, retention time decreases from established 5 rep. Next sample injection, everything back to normal and right in line with the 5 rep. Final injection was of the standard, again, right in line with the 5 rep. So we seem to have one problem injection.
Since I'm new to the combipal scene I wandered what would happen if I injected different amounts of a standard. 0.5, 1.0, and 1.25µL used. Both compounds generated 0.9996 correlation coef. Also, RT's right on top of each other and the tailing was spot on. My thought was what if I had an air bubble that might have reduced my injectable amount. Doesn't seem to be the case.
What direction should I go now?














 
																							