What do you think of this:
I opened 32 ampoules of one of our products and placed them directly in the Acquity autosampler (no transferring to LC-vials). The analysis ran overnight and when I got the results it turns out that all samples were oxidised.
In each row of the autosampler, the first vial is the most oxidised. Then the oxidation decreases for every vial until the last position, no 8, in each row, which is again high. This trend is 100% true for all four rows. The ampoules that are most oxidised are placed closest and furthest away from the cooling fan of the autosampler (10°C).
The problem is that the product is not sensitive to air oxidation in any other place than in the autosampler. If I place opened ampoules on the lab bench for weeks, nothing happens. I have tried to bubble 100% oxygen through the solution, but no oxidation. Any idea why the solution decides to oxidise in the autosampler? Each solution is injected only once, so it has nothing to do with the injection needle.