Hi every body...

PLZ, what is the appropriate temperatures setting in GC for compounds, I mean what is the rule for setting temp. for injector, column and detector. I set this conditions for ethanol:
Injector: 150 c
Column: 90 c
Detector: 190 c
Is this right?

Thanks

Depending on the column, injector and column temperatures look to be OK. Detector temperature depends on the detector.

So, what column and detector? (And have you checked the manual for the instrument for standard detector conditions. Most of manuals for newer instruments can be found on line.)

Thanks for replay

it is FID detector and Packed column

Don is correct.

You could increase all three temperatures to 200°C and still be fine, as long as your column can tolerate that temperature.

I would not go below 150°C for injector or detector in most circumstances.

I would not go below 60°C for the column. But there are exceptions for special circumstances.

Rod
4 posts Page 1 of 1

Who is online

In total there are 17 users online :: 1 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry