Please note that I did not use an injector with a split, the split had been by-passed by the extension of the liner which had a 100% flow into the column.
The Varian autosampler has a variable injection rate which is controlled by the computer.
But to simplify and to accelerate the research I did my injections manually. And did them in a controlled manner (VERY SLOWLY) so the vapor cloud never reached a volume so that it would exceed the flash volume of the liner.
My samples were permitted to refocus on the column and the large amount of solvent permitted to pass through the column before the oven temperature was allowed to increase and move the analyte bands down the column.
Unfortunately, a split/splitless injector is not designed in this manner and its flash volume is too small to inject a large sample without having solvent and analytes going everywhere possible as the volume and the pressure increases during the injection.
Success of this technique is dependent upon the creation of the vapor volume from the sample being less than the amount of vapor entering the column per unit amount of time.
The type of solvent and the analytes, the film thickness of the column and the measured stability of the hand of the chemist are all factors which affect the result.
Use the right hardware and the right technique and improvements of two or even three orders of magnitude can be attained. But if everything is not appropriate, then tailing and wailing are usually the result.
Good luck,
Rod