Hi im currently analysing amino acids in plasma using hplc. I was wondering how to do a calibration curve for the different amino acids in plasma as the plasma itself has amino acids. What type of formula should i use?
-
- Posts: 1680
- Joined: Sat Aug 23, 2008 12:04 am
This depends a bit on what you are trying to do. In most cases, you will add known levels of amino acids to samples of plasma and, if appropriate internal standard or standards. To account for amino acids present, you will need to determine where the response (or relative response to the internal standard) for the amino acids (typically plotted on the Y axis) would be zero. This will be the point at which the the plot (normally a linar plot) crosses the X axis. The value on the X axis would show you how much you need to add all concentrations to move the curve through the intercept. (The vaule read from the X axis is the negative value of the background concentration.) You need to be careful with this calculation it makes a couple of important assumptions: 1) the curve is linear below the lowest level you can measure. 2) the resonse is only due to the analyte of interest.
If your response in the range over which you can acquire data points (don't forget to run a sample into which no analyte was added) is not liner - you have a problem already.
You have to be very careful on how you detect your analytes. In some cases, you may be able to report your analytes only as "not greater than" some the value you determine by reading the result from the calibration - because you can not prove that you do not have an interferece.
If your response in the range over which you can acquire data points (don't forget to run a sample into which no analyte was added) is not liner - you have a problem already.
You have to be very careful on how you detect your analytes. In some cases, you may be able to report your analytes only as "not greater than" some the value you determine by reading the result from the calibration - because you can not prove that you do not have an interferece.
-
- Posts: 7
- Joined: Tue Oct 26, 2010 12:50 am
Hi! ok thanks. this helps a lot. would it be the same as running a calibration in water instead of plasma? i have done this already but would i get the same readings in plasma?
-
- Posts: 1680
- Joined: Sat Aug 23, 2008 12:04 am
It is aways best to check with spiked matrix. There are a number of factors that can affect the respone of your compounds - and unless you check, you can not be sure. And to avoid the question, the general rule (for me, at least) is: if you can build a calibration curve in matrix, do it.
-
- Posts: 7
- Joined: Tue Oct 26, 2010 12:50 am
Ok. Thanks!
5 posts
Page 1 of 1
Who is online
In total there are 8 users online :: 1 registered, 0 hidden and 7 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 7 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 7 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science