Theoretically how is TOF resolving power supposed to change with m/z? Based on observed values for calibrant ions (covering 100-3000 m/z), resolving power appears to increase in something like a logarithmic or power fashion. Also wondering why it seems to plateau out at the high mass end (1000-2000)? This question is asked in the context of small molecules - in case matters.
In mass spectrometry resolving power (
R) is M / Δm, where M is the
m/z value of the ion used in the calculation, i.e.,
m/z 250, and Δm is the difference in the
m/z values of two mass spectral peaks of equal intensity that can be separated so the overlay of the base of the two peaks is 10% of the height (10% valley method). Having such a situation is not always possible and an alternative for delta m is the width of the mass spectral peak at 1/2 of its height (FWHM method). The FWNM method of calculating
R results in a value of
R that is 2x greater than the value obtained with the 10% valley method. Δm is the resolution which is the difference in the
m/z values of two mass spectral peaks that can be separated and is usually reported for a given value of
R. Many people confuse resolution and resolving power.
A linear TOF mass spectrometer has a constant resolving power throughout the
m/z scale. As M gets smaller the difference in the
m/z values of ions that can be separated gets smaller. The ability to separate ions of small differences in
m/z values is better at the low end of the scale than it is at the high end. Most reflectron TOF instruments operate at a constant resolution (Δm), which means that the resolving power (
R) becomes a larger number as you go up the
m/z scale.
Are you asking about a linear or reflectron instrument? The size of the ion has no relevance; therefore, small or large molecules are not important.