Dear Rod,
I do not have access to another GC, however I have a shorter column of exactly the same stationary phase. instead of 50 m it is a 15 m one. To be honest, it is hard to tell whether the same tailing takes place as everything is so clogged.
I could buy a new column, but I thought it would be faster and cheaper to get a few brushes and follow the relatively simple procedure for cleaning the injector and if that does not work, then get a new column of the same or even larger length. What are your thoughts about this?
Thank you.
Dear Peter,
I use 4 mm ID glass frit liners. The inlet temperature is 200 ºC and the splitless time 0.8 min.
I have not opened the bottom nut that would give access to the bottom of my 1077 split/splitless injector, yet. So I think that if I do it, I could take the opportunity to clean the whole thing with the brushes and the three solvents.
http://www.chem.agilent.com/Library/Sup ... a16022.pdf
I am using a 100% dimethylpolysiloxane phase colum 50m x 0.32 mm x 5um. I use nitrogen and the headpressure is 20 psi. It is a bit cumbersome and time-consuming to measure the column flow in my system, so I adjust the headpressure to find a compromise solution between resolution and analysis time.
The solvent used to tail, but I sorted that out by increasing the column flow to its maximum. When the splitless time is over, a massive flow of gas cleans out completely the injector chamber and no tailing occurs.
The tailing peak has to be a derivatisation by-product with MBDSTFA, such as TFA. I am not sure yet. Is it important to find out what it is?
Everything was fine until I did a large injection of EtoAc. It is the only peak tailing, the others seem to be fine, actually hidden under the tail of this peak.