At this moment, I´m working on a new HPLC method for analysis of Silybum Marianum (flavonolignands). I used two mobile phases, 1ml/min, column Kromasil C-18 250x4,5, 20 microlitres.
1- AcN/ Water with buffer ph 2,7 (62%/38%)
Sample solvent with methanol.
2- MeOH/ Water with buffer ph 2,7 (50%/50%)
Sample solvent with methanol and water.
With the first mobile phase I only had one peak, my calibration was good.
But with other mobile phase and eluent for sample, I had two peaks because I have best resolution.
My problem is Why is area peak for a standard is different, is normally that for 50 ppm area peak is 1000 mAU and for 25 ppm area peak is 100 mAU or 50 mAU or 20 mAU. I haven´t reproducibility.
I think that problem is The water in sample, and sample preparation, Why?