At end of last week, GC was acting OK. On Monday, 1st injection of solvent acetone showed nothing. Repeat injection same.
Started troubleshooting:
1. Checked air, He, hydrogen tanks. Replaced septum. Checked parameters: OK. Made sure He flow was registering normal (2.8mL/min) and checked that FID was ON and lighted. Ran test chromatogram to ensure signal reaching integrator (3396 Series II): OK.
2. Turned off instrument and restarted (possible brownout during weekend). Checked parameters: OK. Made sure He flow was registering normal (2.8mL/min) and checked that FID was ON and lighted. Injection still not showing peak.
3. Disconnected outlet of column and dipped end in MeOH to ensure actual He flow: OK.
4. I was using the B detector, so I switched detector PCBs between A & B slots (Yes, I turned off the instrument for the switchout, turned it back on, stabilized, re-checked parameters). Still no peaks. Dropped detector attenuation 5 units. Injection of acetone produced a short, grossly distorted "peak" with plenty of noise at approximately the expected retention time, followed by and merged with a wide blob (~1 min wide), and then a couple of tiny, poorly formed peaks.
5. Switched PCBs back, etc. Still only the ugly little peaks.
I'm thinking that the detector PCB or main PCB may be the problem. I'm helping out a friend in his lab and he doesn't know if the second (A) detector PCB is actually functional. Therefore, switching detector PCBs may not have been a meaningful test.
Thoughts?